Hello, 

We have been using the NI 9219 module to measure our Pancake Load Cell Model # LCF500 and Item # FSH00149. Here is our wiring diagram:

MClayp_0-1680736430436.png

The setup was working well until we sent the load cell off for calibration. Since it came back, we have been unable to read it. We sent it back to Futek, but it checked out fine. When I open NI MAX all I see is noise:

MClayp_1-1680737244061.png

Questions:

1. Does the wiring look right? It's worked in the past so think it should be fine.

2. What is the excitation voltage of pins 3 and 5 of the NI Module supposed to be when nothing is connected? I measured the voltage with a voltmeter and found 0 V with nothing connected. Does a load need to be applied for voltage to be shown on these pins?

 

Any recommendations are appreciated. 

 

Thank you for your help,

Michael

 

0 Kudos
Message 1 of 3
(2,221 Views)
Solution
Accepted by topic author MClayp

You are not configuring the NI9219 to be measure a load cell, instead you're configuring it to read a voltage.

santo_13_0-1680743146795.png

 

If you configure it to measure load cell, you will see an option to configure excitation voltage.

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 2 of 3
(2,210 Views)

Thanks for your help. Ended up finding out that our Python program for reading the load cell was set up to read voltage. I had wired it up as a strain gauge as you pointed out. Just had to rewire it for voltage and supply excitation from a power supply. Thanks

0 Kudos
Message 3 of 3