Hello,
To Brian,
Is your data aquisition time very slow? I'm experiencing the same problem as you described and it seems to be fixed by setting the reset pin to false on DMM Initialize.
To Logan,
I am upgrading my instrument drivers and LabView, but the data aquisition time turned out to be three times slower than the my old setup. I am using a PXI-1052 with doing synchronous triggering with a DMM4060 and a SCXI-1128. Here are my old and new driver and LabView versions:
Old: LabView 6i, NI-DAQ 6.9.3f3, NI-Switch 1.6, NI-DMM 1.6, MAX 2.2,
11-18-2005 10:57 PM
sorry, I accidently hit a wrong button and submitted...
to continue:
Old: LabView 6i, NI-DAQ 6.9.3f3, NI-Switch 1.6, NI-DMM 1.6, MAX 2.2, VISA 2.5.2
New: LabView 7.1, NI-DAQmx 8.0, Traditional DAQ 7.4.1, NI-Switch 3.0, NI-DMM 2.5, MAX 4.0, VISA 3.4.1 (these are the latest as I'm writing this)
Thanks to Brian finding out about the reset pin.
Also, I couldn't use the test VIs Logan put up. It's missing Digits_to_AbsRes.vi, where can I find it?
Here is my email if any of you would like to contact me: leon.y.su@dupont.com
Thanks,
Leon