annuler
Affichage des résultats de
Rechercher plutôt
Vouliez-vous dire :
use of 9219 module for 4 RTD measurement with excitation current in series
Résolu !
Accéder à la solution
Is it possible to use the 9219 module for the measurement of 4 RTDs using a single excitation current through the 4 resistors in series and measuring voltage with the 4 channels as on the appended sketch.
I am trying to limit the number of lines as space on the PCB is limited
0 Compliments
Message 1 sur 5
3 269 Visites
Accepté par DidierM
Hello.
This connection is possible, except that the current will be generated by one of the 4 channels, between EX+ and EX-.
Best regards
0 Compliments
Message 2 sur 5
3 243 Visites
I was wondering that the module would not accept this configuration.
For the CH1, configured in 4W resistance,  it will be as if the wire resistance is 3 times the measured resistance.
CH2-4 will be configured in simple voltage measurement (value of resistance is obtained by dividing by the current of CH1)
Am I correct ?
Regards
Didier Maillefer
0 Compliments
Message 3 sur 5
3 240 Visites
Hello.
You're right.
The first measurement will be a 4W with a wire resistance equal to 3 times the value to measure, and the 3 other measurements will be a 2W.
Best regards.
0 Compliments
Message 4 sur 5
3 222 Visites
Thank you for your valuable help.
I have received the 9219 module from NI and started to experiment.
The measurement does not work the way I expected.
I have configured the first measurement (CH0) in 4W resistance and the three others (CH1-3) in voltage.
You tell me to configure CH1-3 in 2W resistance but then, they will have their own current source and I will lose the benefit of the 4W configuration.
I can measure the voltage on CH1-3 but it does not correspond to the value of the resistance given the excitation current set at 500uA.  When I measure the actual excitation current through the 4 resistors, I read 396uA and it varies with the measured resistance.
I have tried to use the property node giving the AI.Excit.ActualVal thinking that it was a measured value of the actual current but it outputs the theoretical value of 500uA.
At this point, I can't figure out how to obtain an accurate resistance value on Ch1-3
Can you help me?
0 Compliments
Message 5 sur 5
3 174 Visites