Hi all,
I realised that a trigger rate of 100kHz is not sufficient for my application, I need at least a few hundred Hz. I was wondering if I can change my analog triggering wave (more than 100khz) into a TTL signal externally and input the TTL into the digitizer. Will this actually enable me to achieve a higher trigger rate or am I still bounded by the re-arm time?
Sorry if it does not makes sense.
Regards,
lfw
01-05-2007 08:42 AM
lfw,
The same trigger rearm time applies, regardless of the trigger source.
Is your trigger always periodic? If so, could you trigger once and use the deep onboard memory of the 5922 to store the entire waveform, then fetch back only the portions you are interested in?
02-09-2007 03:40 AM
02-09-2007 10:51 AM
Hi Andy,
From your word document, it looks like your trigger occurs immediately after each acquired record. What might be happening is that your trigger condition for the next record sometimes occurs within the previous record's acquisition, so the digitizer misses that trigger. You can test this by reducing your record size. Try using 90,000 instead of 100,000. Do you still miss triggers?
If this is the case, then your best solution would be to reduce your record size to ensure that you are not still acquiring data when the triggers occur. Or simply acquire all the data in one waveform (as in the second graph) and split it into 10 waveforms in software.
I hope this helps. Let me know if you have any further questions.
Regards,
Sean Close
Product Support Engineer: High-Speed Digitizers
02-09-2007 02:05 PM
Hello Sean,
Thanks for the quick response.
It's been quite long just to figure out this matter. :s
I'm wondering what the real definition for trigger is. I didn't use the 'real' triggers as the curve will come every 10 ms by setting a frequency of 100 Hz. Can this 100 Hz frequency be called as trigger as well?.... just wondering if we are in the same way of thinking 🙂
In general, if we set the number of record of 10, will the program read the data after each trigger occurs, even we don't have 'real' triggers (only a signal which has a certain frequency)? For example, when we have a sine wave signal with frequency of 100 Hz, and want to read the first 5000 points of data with sample rate of 10 MS/s, then the program will read exactly 5000 points every 10 ms after the first reading. (without any missing triggers).
Hope you are not confused with my 'story', as I'm not really good in explaining technically.
regards,
andy
02-09-2007 02:27 PM
Hi kentang,
"I'm wondering what the real definition for trigger is. I didn't use the 'real' triggers as the curve will come every 10 ms by setting a frequency of 100 Hz. Can this 100 Hz frequency be called as trigger as well?.... just wondering if we are in the same way of thinking" I was assuming that you were triggering off of the same signal that you were acquiring. In this case, triggering everytime the signal crossed the x axis with a rising edge. Is this not the case? If not, then what was your trigger signal?
02-09-2007 02:44 PM
02-09-2007 02:56 PM
02-09-2007 03:34 PM
02-09-2007 03:41 PM
Hi kentang,
If you only take 1000 samples per record and you acquiring data at 10 MS/s then each record will span 100 us. If your triggers occur every 10 ms, then you should not miss any. I believe you were missing triggers before because you were still acquiring data from a previous trigger when a new trigger occurred. You could test this be reducing your record size.
Let me know your results.
Regards,
Sean Close