Hi Vsh,
I'm sorry but the 5922 does not support exporting the internal sample clock to the CLK OUT connector.
The manual [1] has a routing diagram which shows what signals can be routed to the connector, I've placed it here
for reference as Figure 1 below.
Fig 1. Clock Routing Diagram
You can route the CLK IN external connector to the CLK OUT external connector, as the diagram shows.
Figure 2 shows how to do this, these are all niSCOPE palette vi's.
Fig. 2 CLK IN to CLK OUT
And finally, here is what will throw an error with the NI 5922, as seen in Figure 3 below.
Fig 3. Error 89136
What options do you have?
#1. Assuming this is to synchronize the digitizer with something else, then use a external
device's sample clock.
#2. Route the PXI CLK to the CLK Out terminal and use this for sample clock source/synchronization
#3.
[1]