03-04-2015 05:14 PM
For instance when using the following commands
niScope_ConfigureDigitial (with a positive delay time)
NiScope_SetAttributeViInt32 ( NISCOPE_ATTR_FETCH_RELATIVE_TO to NISCOPE_VAL_READ_POINTER)
followed by
niScope_FetchBinary16 (looped to retrieve data plus wfmInfo)
What will the value of the wfmInfo.relativeInitialX prior to the external trigger? Is it defined? Is it set to a fixed number to indicate triggering has not occurred until the trigger has been received and then it becomes accurate after the trigger has been received?
I would like to use this to do a modified FecthForever but synchronized to an external trigger for waveforms up to about a minute.
It seems the trigger delay for the 5122 should be able to hande a minimum trigger delay of about 170 seconds (at max sample rate more at lower rates), so the data I would need could all be borrowed from the prefetch buffers of the externally triggered and properly post trigger delayed measurement.
Thanks in advance for any advice.
Greg
03-10-2015 02:57 PM
In the case that you are fetching pre-trigger data (fetching data before a trigger has occured., relativeInitialX will be the same value as absoluteInitialX.
Nathan
03-12-2015 04:04 PM - edited 03-12-2015 04:04 PM
You are exactly correct. I will use the transition from equality to inequality to signify the physical trigger location.