06-07-2007 08:09 AM
06-08-2007 07:30 AM
06-11-2007 08:14 AM
Thanks for the message. The board I am having problems with at the moment is revision 'E'. It is running with NI-SCOPE Version 2.6, NI-DAQ Version 7.1, and is run on Lab Windows/CVI Version 6.0 under Windows 2000. According to your reply this board should calibrate o.k with these software versions.
06-13-2007 03:28 AM
Hello Stu,
Does this particular board show this problem on all versions of software that you run or is this still part of the batch where you believe the behaviour is different in different versions?
If it does, then I would suggest that you call into your local NI office and speak to someone about this - if it has been returned for repair before, then if you have those details handy it will be helpful.
How many boards are you seeing this problem with?
Regards
Hannah
NIUK & Ireland
06-18-2007 08:07 AM
The board we are having problems with at the moment will self-calibrate if put into a machine that has NI-Scope version 2.0 installed with NI-Daq 6.9.2. But not with any later version. However we have another PXI-5112 which has been sent away for repair recently for same problem and have been advised that the relays are faulty. Could this board have the same problem?
Best Regards,
Stuart Brydges
06-20-2007 05:51 AM
Hi Stuart,
What other versions of the NI-DAQ and NI-Scope software have you tried with this card? Are there any other versions - earlier or later - that will allow you to self calibrate this card.
Looking at the next version up from 6.9.2 - which is 6.9.3 - and this doesn't show any changes in the driver which could cause this with the 5112.
It could be a similar problem to the board you sent off for repair, but if you can calibrate with multiple versions of the software then it does not seem likely
Regards
Hannah
NIUK & Ireland
06-20-2007 06:45 AM
Hi Hannah,
The only other versions we have are, NI-Scope version 2.1.4 with NI-DAQ version 6.9.3. and NI-Scope version 2.6 with NI-DAQ version 7.1. most machines are running NI-Scope version 2.1.4 as they have Lab/Windows CVI 5.5 on them under Win98.
None of these versions will calibrate the card.
06-20-2007 08:03 AM
Hello Stuart,
I believe you are currently dealing with one of my colleagues Lee over the telephone about this issue. I would recommend that you continue to work with him on this issue for now, as we will perhaps need to get the cards sent into NI UK for some diagnostics, which will require us to find out serial number information and direct contact details, which are perhaps not best discussed in the forum.
Regards
Hannah
NIUK & Ireland
06-20-2007 10:50 AM
06-21-2007 03:36 AM
Hi Kurt,
I have read the KB and it says to me that the PCI version of the NI-5112 was released at the same time as revision 'E' of the PXI version of the NI-5112 and will work with NI-DAQ version 6.9.0 or later. I believe card 2 to be faulty also, so I will arrange return of this card also. As for the PXI versions we have it looks like these have faulty relays so will arrange repair of these too.
Many thanks for your help,
Stuart Brydges