10-22-2014 06:49 AM - edited 10-22-2014 06:49 AM
Every one.
I am using Labview 2012. I am trying to interface Keithley 2401 SMU with laptop via Keithley-GPIB KUSB 488B. When I am trying to run the program, it is indicating error 1073807202. I have reinstalled all the labview software, NI-VISA (latest), drivers and 2401 library files, but the problem remain as the same.
Can any one hepl me. I need to fix this problem as soon as possible as I need this for my research purpose.
Also can u suggest me the SOP (Standard operating procedure) sequence for installing labview, NI-VISA and instrument driver.
Thanking you
Avishek Das
10-22-2014 07:51 AM
10-21-2015 03:42 PM - edited 10-21-2015 03:43 PM
Have you solved the problem? I use the same keithley 2401-kusb-488b with the same error. I can;t run any of the VIs. Have you tried with rs232?
06-10-2016 09:56 AM
Hello,
I have the same error code. Could you solve it? Please can you help me?
06-13-2016 06:44 AM
Hey Everyone,
Since we are working with a GPIB based instrument, you are also required to install NI 488.2 drivers. This is different from NI VISA driver.
Another major cause of this issue if the VISA driver gets corrupted during installation.
Please refer to the Knowledge base article mentioned below for detailed description on install instructions of VISA drivers.
http://digital.ni.com/public.nsf/allkb/10C6C0FBB8E7BD4486257850006CE8FF
Also, another issue that was reported on similar lines is mentioned in this forum post
Thanks & Regards,
Raghunandan N V
AE - NI India