06-14-2016 10:22 PM
I'm trying to read oscilloscope waveform through LabVIEW but when I read the waveform, instead of actual voltage, numbers ranging from 0 to ~2000 is shown in LabVIEW VI. When I vary the position of the waveform on the oscilloscope, these numbers also change. Please check the below image. Is there any settings that I am missing? The voltages on the scope are around 5 volt. Please help.
Solved! Go to Solution.
06-15-2016 01:04 AM
Are you using an Agilent instrument driver subVI to get this waveform? If yes, can you please tell us what driver (model/version) are you using and what's the name of the subVI that produced this waveform? I'll have to download the drivers and take a look inside that VI.
06-15-2016 01:15 AM
Hi,
Yes, I'm using Agilent 6000 series drivers for LabVIEW (Driver page). I am using LabVIEW 2015. I'm using "Read Multiple Waveforms.vi" to get the waveforms.
06-15-2016 10:30 AM
Hi,
You're using the wrong driver.
Try
http://sine.ni.com/apps/utf8/niid_web_display.model_page?p_model_id=10028
Curt
06-15-2016 11:14 PM
It worked; I'm able to see proper values. Thanks!