05-14-2008 01:50 AM
05-14-2008 06:39 AM
Hi,
Did you install the LabVIEW compatible drivers as well.
Check with Keithley on their website and make sure you have the latest drivers.
Regards
Ray Farmer
05-14-2008 05:57 PM
05-15-2008 02:05 AM
05-21-2008 08:30 AM - edited 05-21-2008 08:32 AM
05-28-2008 01:42 PM
The version 8.0 of the CEC488 driver you have is quite new and should do the job (yes there is a version 8.3 of this driver which supports your card, but the only difference is added support for new KUSB-GPIB devices).
It does not make sense to attempt to communicate with your instruments using the Agilent Connection Expert, instead...remove it and any other GPIB drivers. If in doubt, remove and reinstall the CEC488 driver.
You should first use the CEC TEST488 Utility to make sure your card is installed and properly recognized. If all goes well, then run the CEC TRTEST utility to communicate with your instruments via GPIB directly. If all goes well, you can then use the NI Visa Interactive Control to talk to instruments via VISA (if you have problems check your VISA passport. There are both KeViCEC and NI GPIB passports. Turn off the NI passport and leave the KEViCEC marked on and try to communicate again).
This should allow you two ways to use LabVIEW to communicate with your instruments:
1.Use VISA commands or
2. Use CEC driver supplied GPIB commands