12-16-2014 01:16 AM
Hi,
I am using a GPIB-USB-HS on my PC and i wan tto connect Agilent B1500A instrument with it. But when i run MAX, it says that insturmen tis not found.
Could you please help me out with it.
Thanks
Rajesh Agarwal
12-16-2014 01:52 AM
12-16-2014 04:36 AM - edited 12-16-2014 04:37 AM
Thank you sir for your reply:
When i connect the GPIB to Agilent B1500A, MAX shows that it is connected.
Regards,
12-16-2014 05:20 AM
Do I need to install Agilent I/O library in my PC also.
Thanks
12-16-2014 06:30 AM
12-16-2014 12:47 PM
Sorry for the confusion.
My PC MAX shows that there is a GPIB hardware (Image attached), but my instument (B1500A) which i want to control thorugh GPIB, its MAX shows a question mark on GPIB(image attached), but the Agilent I/O library shows that a GPIB is connected (image attached). When i scan my PC for GPIB connetced devices, it shows that the instrument is not found (image attached)
I had installed the latest verison of VISA And MAX on my PC and the instrument. I had also disabled the Agilent 488 properties on the instrument.
I had set the VISA interface ID as GPIB0 and address as 8 on both my PC (MAX) and the instrument (Agilent I/O library, because MAX shows question mark under Devices and interface.
My PC is 32 bit XP and the instrument is 32 bit windows 7.
I am very much upset with this problem as my supervisior has told me to interface this instrument with LABVIEW, but i am not able to even connect the instrument with PC. Please help me out.
Thanks.
12-16-2014 01:35 PM
12-16-2014 04:03 PM
If you are connecting the B1500A to your PC you don't have to do anything with MAX on the instrument itself.
The B1500 you mentioned has an OS - many times Agilent/Keysight instruments that have an OS have a separate application where you can configure the remote interface. You will have to check the manual/s to see how to configure the remote interface and the recommended settings. http://www.keysight.com/en/pd-582565-pn-B1500A/semiconductor-device-analyzer?&cc=US&lc=eng
01-04-2016 06:39 PM
B1500A mainframe has its own Keysight/Agilent GPIB system controller. Try following these steps.....
On B1500A Mainframe…..
On Host PC…..
On B1500A Mainframe…..
On Host PC…..
On B1500 Mainframe…..
Hope this helps...
01-25-2017 05:58 AM - edited 01-25-2017 05:58 AM
To use the instrument remotely, some things have to be done, because the instrument is prepared to work with their own interface EASY EXPERT.
If the EasyEXPERT is running the remote control will not be activated, the first thing is exit the EasyEXPERT. Then the first time the GPIB connection has to be configured following the next steps: