Hello.
I have a question about the CAUTION in HP 4155A instrument.
We have a problem to control the instrument because of the error in the picture below.
Before the error occurs we used the instrument well by GPIB0 communication.
Solved! Go to Solution.
09-24-2020 08:43 AM
No error attached.
Can you perform the same task manually?
I suspect you have not setup the interlock properly. See - https://forums.ni.com/t5/Instrument-Control-GPIB-Serial/HP4155A-Fixture-Open/m-p/4056371#M81533
Craig
09-24-2020 09:13 AM
Oops, thank you for your comment.
I attach the picture of the error.
09-24-2020 09:06 PM
Yikes, that's not an operation error but an SMU error from the calibration/diagnosis page. I expect it appears even if you cycle the power?
Looks like the SMU has been damaged and needs repair. You should contact Keysight for support.
Good luck.
Craig
09-25-2020 09:46 AM
Yes, it appeared right after when I turn on the power.
Thank you for your help.
I'll contact Keysight for support.
Thank you so much.
Jea Jung Lee.