12-15-2015 07:38 AM
Hello everyone,
I’ve encountered an issue with using the Basler camera (Basler ace acA2040-180km NIR). For some reason I am not able to establish/maintain the serial connection to the camera using e.g. Pylon viewer. Other than that, the camera works fine and I can acquire data as usual, however I cannot change the parameters like exposure time or area of interest.
Used Hardware & Software
Hardware
Software
Where it all started
The whole imaging system is connected to an isolation transformer as a safety precaution. During standard image acquisition, power went off in the lab thus shutting everything down. After the power came back online, I started the computer and wasn’t able to establish a serial communication to the camera anymore (using Pylon viewer from Basler).
Troubleshooting so far
When I use the Pylon CL Configurator, it detects the FlexRIO port but gives a message: “The stream is not a CLPort registry stream since it is missing the magic GUID in the first line”. And as an intermediate user I have no idea what the message is about.
Long story short
The serial communication between the PC and the camera is not working and for some reason Pylon viewer is not able to detect the camera. I’m so baffled myself why this is not working so if someone can shed light on this mystery I'd be grateful.
Best,
Kari Vienola
Solved! Go to Solution.
12-16-2015 08:16 AM
Hello Kari,
Thanks for posting so much information about your problem. I've not been able to find any documentation about that error code, but having duplicate systems is very helpful. I may have misunderstood but it sounds like you've been able to establish that the camera is not the problem. Have you tried the PCI card in a different slot in the PC? Additionally, if you move the PCI card, does the problem migrate or is it resolved?
Best regards,
12-18-2015 10:22 AM
Hello Jake,
Yeah currently we have done several things:
What I can deduct from this troubleshooting is that the frame grabber, cables or the camera is not the origin of this issue. The issue lies in the software but pinpointing it out is like finding a needle in the haystack if it's something like a corrupted dll-file somewhere (or at least with my skillset it is).
I will spend few more days investigating the issue but currently the only option seems to be formatting the OS drive and reinstalling everything. If you have any further suggestions, I'll be glad to test them out.
Best,
Kari
03-01-2017 08:12 AM
Hello Kari:
I also have the same problem and the same reason.
Have you solved the problem and how?
Thank you very much.
03-02-2017 09:06 AM
Is the serial bus transformer connected? Try to temporarily by-pass the transformer and check for communication. If communication works, then the transformer may have an issue. You might want to try optocouplers for isolation on the serial bus rather than the transformer. A transformer will pass along spikes to the other side (di/dt) where as an optocoupler will clip any spikes. Also, a discontinuity in the transformer core may cause odd behavior.
03-05-2017 08:40 AM
Thank you , the problem have been fixed by follow steps:
When you can't find camera link camera in pylon viewer, and the pylon CL configurator say "The stream is not a CLPort registry stream since it is missing the magic GUID inthe first line", you can solve the problem like below.
Cause: A file may not have been created correctly during installation or may have been corrupted/overwritten at some point by other 3rd party software.
Solution: 1. Delete the PortIDDeviceIDCache.ini or PortIDDeviceIDCache2.ini file, found in the folder below.
2. Relaunch the CL Configurator and probe the ports again to generate a new .ini file.
(Please note that these folders may be hidden by the operating system)
pylon 2.3 (PortIDDeviceIDCache.ini)
C:\ProgramData\genicam\xml\cache\
pylon 3 (PortIDDeviceIDCache.ini) and pylon 4 (PortIDDeviceIDCache2.ini):
(WIN7)
English: C:\Users\All Users\genicam\xml\cache
German: C:\Benutzer\All Users\genicam\xml\cache (WinXP)
English: C:\document and settings\All Users\application data\genicam\xml\cache\
German: C:\Dokumente und Einstellungen\All Users\Anwendungsdaten\genicam\xml\cache\
Other problem:http://www.baslerweb.com/en/support/knowledge-base/frequently-asked-questions?page=5