Assuming you are using GPIB, one solution is use multiple listener operation. Although you normally send a device message to a single listener, you can also send it to multiple listeners at a time. How to do it is different depending on what library you are using.
If using NI-488.2 API from LabVIEW, you can use GPIB 488.2 "SendList" function that accepts multiple listeners through the "address list" parameter.
If using NI-VISA, you need open a separate session with "GPIB0::INTFC" then use VISA GPIB Command function (for ATN-TRUE part, specifying multiple listeners in it) and VISA write function (for ATN-FALSE part) to construct a multi-listener operation.
07-03-2006 02:15 AM
07-03-2006 02:15 AM
07-05-2006 04:22 AM
07-05-2006 08:27 AM
OK, I have found some more information about these multimeters and the phenomenon that puzzled me. It seems that the problem wasn't triggering at all, but the integration time. I now monitor the EXTOUT APER signal (TTL) which shows the integration period.
In these multimeters you can set the integration time of the measurement using NPLC command, 1 NPLC = 1/50 of a second (if used 50 Hz AC supply as it is in Europe). And when I set the NPLC numbers equal to both multimeters (I experimented numbers from 10 to 100) , the end of the measurement cycle was longer for the current (DCI) measuring multimeter. The time difference was exactly 2, ie. current measurement took 2*NPLC*(1/50) seconds and the voltage (DCV) measurement took as it supposed to be, NPLC*(1/50) sec.
But it got more weird. I usually cofigure the multimeters using LabView (GPIB bus). But when setting the NPLC form the display unit (let it be for exampe 50) and then measuring voltage (DCV), then the integration time is twice longer as it supposed to be, ie 2*NPLC*(1/50) seconds. Do I leave something important out what to cofigure or is there anything special about these multimeters?? What concerns about the AUTORANGE, then this is turned off and the range is set manually. The short table below gives an overview how these multimeters work at the moment.
NPLC command set via GPIB ------------------------ NPLC command set using the keyboard of the display unit
DCV measurment integration time NPLC*(1/50) sec 2*NPLC*(1/50) sec
DCI measurment integration time 2*NPLC*(1/50) sec 2*NPLC*(1/50) sec
Meelis
07-06-2006 05:42 AM
03-07-2008 12:23 PM
02-09-2011 02:16 AM
Hi Makoto.
I like your answer for simultaneous triggering.
I'm also working with two HP3458As.
I'm intrested how to work with VISA GPIB Control ATN ---> GPIB0::INTFC.
I put it between two GPIB Write functions.
If my first HP is connected two PC, I supose that I don't need to use VISA GPIB Control ATN before writing to it.
Is this true?
Kristina.
09-14-2016 05:27 PM
HI , I am new user using HP3458a and like to what commands i can use ( in GIPB) to set 2 HP3458A at the same time to make DC voltage measurement from 2 meters. I am trying to cut down the Trigger time so like to know if I can trigger
both at the same time. so I cna save trigger time if done in parallel.