08-31-2012 12:36 PM
Now you've disable the controller from sending EOI?
The EOS byte does not look right. It's a line feed which is 0xA. I don't remember if that field is hex or decimal.
11-23-2012 02:57 PM
Thanks Woodster.
This also fixed my problem, it says timeout error, but it is really conflicts between dlls.
BR,
JMA
10-19-2016 01:38 PM
Hi all
I am using the NI VISA I/0 to talk to Keithley SMU 2410 using a PROLOGIX GPIB-USB Controller.
I have correctly configured(I think) the SMU to VISA using the GPIB connection but I'm getting the same timeout error as mentioned in this thread.
The command outputs are the following:
_____________________________________________________________________
Command: *IDN?
Write: Write Operation (*IDN?)
No Error
Read: Read Operation
Return Count: 82 bytes
KEITHLEY\sINSTRUMENTS\sINC.,MODEL\s2410,4116375,
C33\s\s\sMar\s31\s2015\s09:32:39/A02\s\s/J/M\n
____________________________________________________________________
After this, I try to use the following command sequence for measuring current by sourcing voltage:
*RST
SOUR:FUNC VOLT
SOUR:VOLT:MODE FIX
SOUR:VOLT:RANG 0.2
SOUR:VOLT:LEV 0
SENS:FUNC "CURR"
SENS:CURR: PROT 100e-3
SENS:CURR:RANG 10e-3
OUTP ON
READ?
The output I get after I read each of these commands is:
Read Operation
Error
VISA: (Hex 0xBFFF0015) Timeout expired before operation completed.
I am struggling to understand why the I/O is not able to souce voltage on the sourcemeter even when it responds well to the *IDN? command.
Let me know if any more information is needed.
Thanks!