I'd like to measure my experiment data by using two Keithley 2000 which are controlled by Labview software.
But I have some problem this labview block diagram, that I wrote, it can run but the sequence of channel changes every time, and both keithley 2000 aren't simultaneous.
I've connected each Keithley 2000 together by hp10833A, as attach picture file.
If who do have any idea to write Labview block diagram better than me, please introduce me.
Thank you.
07-08-2020 05:38 AM
Hello Everyone,
I have the same problem as WANISON. Is there anybody, woh can help me?
Thank You !
07-08-2020 05:51 AM
Dear FlorianDirnberger
Please try to set "Delay" for each channel as slow as possible.
primary setting is 0 s, try 0.1 or more.
Please see at Configure Trigger.vi of KE2000 driver for Labview.
Best Regards,
07-08-2020 06:04 AM
Hey WANSION,
thanks for your help. Do speak german? Then i could ask you my questions more detailed.
07-08-2020 06:50 AM
Dear FlorianDirnberger
Sorry, I can speak English and Thai.
Best Regards,
07-09-2020 03:25 AM
Dear Wanison,
it doesn´t work with the Time Delay. How are your settings for "sample count" , "scan mode" , "number of readings" etc... Maybe i have to adjust different settings in the programm or in de Keithley 2000. I hope you understand what I mean and can help. Thanks.
07-15-2020 09:39 PM
Dear lorianDirnberger
For sample count set as same with primary setting.
Scan mode: Internal
Number of readings : depend on how many channel you use.
For example;
Starting channel = 1
Stopping channel = 8
Number of readings = 8
I hope this can help you