Analog signal: wrong reading in the test panel
Dear All,
I am writing since I am facing a quite strange problem.
I connect a sensor to my DAQ. If I read with the tension at the terminals on my DAQ with a multimeter, I read 3 V (and that is the correct value).
The point is that on the test panel I read a smaller value (1,2 V) and I can not understand why.
I also tried to use a different analog input, but nothing change.
Samuele
(568 Views)
What DAQ?  How is the sensor wired to your DAQ?  How is the channel configured?
GCentral
(562 Views)
Sorry, I forgot some details:
DAQ: USB-6212
Sensor: data sheet attached
Channel configuration like in the picture.
(523 Views)
Are both sides of the differential signal already referenced to the AI GND?  If they are floating, you need to add a couple of large resistors (1MOhm usually works) on each side to AI GND.
For more information, read this article: Field Wiring and Noise Considerations for Analog Signals
GCentral
(511 Views)
You have not shared any connection diagram or information about your sensor.
Santhosh
Soliton Technologies
(495 Views)
Sorry,
I forgot the data sheet.
(472 Views)
Attached is the English version of the datasheet.
Now I see your issue on page 6 of the datasheet.  This sensor's output is not a simple voltage; it outputs a 4-20mA signal.  What this means is that it will output 4mA for the lower range and 20mA for the higher range.  You cannot directly measure voltage because its output is a current.  A simple way to handle this is to put a 500 Ohm resistor across your DAQ inputs.  This will allow you to use the 10V scale for the DAQ.
20mA*500 Ohm = 10V
4mA*500 Ohm = 2V
GCentral
(459 Views)
