Dear All,
I am writing since I am facing a quite strange problem.
I connect a sensor to my DAQ. If I read with the tension at the terminals on my DAQ with a multimeter, I read 3 V (and that is the correct value).
The point is that on the test panel I read a smaller value (1,2 V) and I can not understand why.
I also tried to use a different analog input, but nothing change.
Thanks,
Samuele
01-15-2024 08:15 AM
What DAQ? How is the sensor wired to your DAQ? How is the channel configured?
01-15-2024 09:46 AM
Sorry, I forgot some details:
DAQ: USB-6212
Sensor: data sheet attached
Channel configuration like in the picture.
Thanks
01-15-2024 11:08 AM
Are both sides of the differential signal already referenced to the AI GND? If they are floating, you need to add a couple of large resistors (1MOhm usually works) on each side to AI GND.
For more information, read this article: Field Wiring and Noise Considerations for Analog Signals
01-15-2024 12:20 PM
You have not shared any connection diagram or information about your sensor.
01-16-2024 02:09 AM
Sorry,
I forgot the data sheet.
01-16-2024 06:54 AM
Attached is the English version of the datasheet.
Now I see your issue on page 6 of the datasheet. This sensor's output is not a simple voltage; it outputs a 4-20mA signal. What this means is that it will output 4mA for the lower range and 20mA for the higher range. You cannot directly measure voltage because its output is a current. A simple way to handle this is to put a 500 Ohm resistor across your DAQ inputs. This will allow you to use the 10V scale for the DAQ.
20mA*500 Ohm = 10V
4mA*500 Ohm = 2V