Dear All,

 

I am writing since I am facing a quite strange problem.

 

I connect a sensor to my DAQ. If I read with the tension at the terminals on my DAQ with a multimeter, I read 3 V (and that is the correct value).

 

The point is that on the test panel I read a smaller value (1,2 V) and I can not understand why.

 

I also tried to use a different analog input, but nothing change.

 



Thanks,

Samuele

0 Kudos
Message 1 of 7
(565 Views)

What DAQ?  How is the sensor wired to your DAQ?  How is the channel configured?


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 2 of 7
(559 Views)

Sorry, I forgot some details:

DAQ: USB-6212

Sensor: data sheet attached

Channel configuration like in the picture.

 

Thanks

0 Kudos
Message 3 of 7
(520 Views)

Are both sides of the differential signal already referenced to the AI GND?  If they are floating, you need to add a couple of large resistors (1MOhm usually works) on each side to AI GND.

 

For more information, read this article: Field Wiring and Noise Considerations for Analog Signals 


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 4 of 7
(508 Views)

You have not shared any connection diagram or information about your sensor.

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 5 of 7
(492 Views)

Sorry,

 

I forgot the data sheet.

 

 

0 Kudos
Message 6 of 7
(469 Views)

Attached is the English version of the datasheet.

 

Now I see your issue on page 6 of the datasheet.  This sensor's output is not a simple voltage; it outputs a 4-20mA signal.  What this means is that it will output 4mA for the lower range and 20mA for the higher range.  You cannot directly measure voltage because its output is a current.  A simple way to handle this is to put a 500 Ohm resistor across your DAQ inputs.  This will allow you to use the 10V scale for the DAQ.

20mA*500 Ohm = 10V

4mA*500 Ohm = 2V

 


GCentral