I found many articles about serial port but all of them are Regarding about RS232 and Rs485 , But i cant found any articles about RS422 serial communication with Labview
By The way i followed the same way and create a VI in LabView But while Executing That VI it shows an error like
""
Possible reason(s):
LabVIEW: An input parameter is invalid. For example if the input is a path, the path might contain a character not allowed by the OS such as ? or @.
=========================
NI-488: Command requires GPIB Controller to be Controller-In-Charge.""
can any one please help me .???
Solved! Go to Solution.
02-13-2017 09:14 AM - edited 02-13-2017 09:14 AM
Hi arun,
create a VI in LabView But while Executing That VI it shows an error like
Ok, so there is an error in YOUR VI.
As long as you hide your VI from us you will have to debug it on your own!
Can we use RS422 serial port communication
As long as you get a COM port in the device manager you will be able to use VISA functions with it…
02-13-2017 09:20 AM
Thank you for responding , i dont have any problem to show my code please take a look on the following attachment
i installed all the drivers and NI-visa also , And in the NI-MAX detect my port as COM23
02-13-2017 09:22 AM
Regarding RS-422, as long as it shows up as a COM port, it will code exactly the same as RS-232.
02-13-2017 09:25 AM
Yeah , it detected as COM port, Thats why i followed the procedure but at the end while executing the code it shows errors as i mentioned in the code
The attachment consist of my VI please take a look
02-13-2017 09:30 AM
You don't have any of the error wires connected. Which function or subVI is throwing the error? Do you have the com port correctly entered? I don't see any com port saved as default in your VISA control.
You have a major flaw in your program in that you use the Bytes at Port method, and you don't have any wait function between the writing of your command and the reading of the response. Your Bytes at Port is likely zero because you haven't given any time for the device to respond with a message.
02-13-2017 09:37 AM
What device are you trying to talk with (make and model)? Based on the error description you gave, I would think you did not specify the VISA session correctly before running the VI. Even then, what is teh baud rate supposed to be? What is the message protocol?
02-13-2017 09:50 AM
02-14-2017 12:16 AM
@ranvensFan i installed everything perfectly and wires are connected according to the tutorial
please look at the following screenshots may you get my problem exactly
02-14-2017 12:18 AM
just look at the attachment , you may get what is the exact problem i am facing @crossrulz