It sounds like you are trying to use the traditional DAQ functions instead of DAQmx. Use the Measurement I/O>NI-DAQmx palette. The obsolete traditional functions do not work with an M series DAQ device.
03-19-2010 02:28 PM
I am facing a similiar problem now, the difference is that I am using PCI-6251 and LabView 7.1. The vi was created using LabView 6.
MAX configured PCI-6251, but I still get the error message below:
Error -10401 occurred at AO Group Config
NI-DAQ LV: The specified device is not a National Instruments product, the driver does not support the device (for example, the driver was released before the device was supported), or the device has not been configured using the Measurement & Automation Explorer.
Anybody can help me? Thanks.
03-19-2010 02:44 PM
03-19-2010 02:52 PM
The VI is using NI-DAQ 7.2.
LabView 7.1 (and previous LabView version?) is not supported by NI-DAQmx.
03-19-2010 02:53 PM
03-19-2010 02:57 PM
Can PCI-6251 use the traditional DAQ too?
I want to repoduce the VI used with the traditional DAQ, don't want to rewrite the program.
Thanks.
03-19-2010 02:58 PM
03-19-2010 03:00 PM
03-19-2010 03:07 PM
Thanks for your reply.
I am using 6251 now.
Do you mean 6251 and LabView 7.1 both work with traditional DAQ and DAQmx?
What setup do I need to make the old VI created by LabView 6 work?
Thanks again.