Hi,
I am new to LabVIEW and I am trying to Generate a sine wave using AO and trying to read it using the analog input. However, when I change the amplitude of the signal while running, the aquired signal does not change (update) immediately with there is a delay of few seconds and I want to avoid it. I am posting my code below
Solved! Go to Solution.
12-01-2023 11:43 AM
Please upload your code saved in an earlier version of LabVIEW.
The latency is device-dependent. What is your device model used for both AI and AO?
12-01-2023 12:35 PM
Hi ZYOng,
I am using NI PXIe- 6345 I/o module there were no latency for single sample acquisition, which LabVIEW version are you using?
12-01-2023 01:20 PM
There is no latency for a single sample, but if you are trying to read or write multiple samples, DAQmx will wait until all data is ready.
From the snippet, the sample rate is 1000. The default sampling info for the function generator is 1000/1000. It will take 1 second for all data to be sent out.
If you want a low-latency AI-AO control, you should use NI-DAQmx Hardware-Timed Single Point (See Case 1) along with the combination of point by point VI, AC & DC Estimator PtByPt.vi and Sine Wave PtByPt.vi