Hello,
I am new with Labview and I am currently trying to connect a Lecroy HDO6034 oscilloscope via USB to perform a data acquisition using a NI driver.
I can see in NI MAX that the scope is connected to the computer but when I try to use a driver I get this error message :
Error -1073807339 occurred at VISA Clear in LeCroy Wave Series.lvlib:Initialize.vi->LeCroy Wave Series Acquire Waveform.vi
Possible reason(s):
VISA: (Hex 0xBFFF0015) Timeout expired before operation completed.
I looked it up online but it did not help me to resolve the problem.
Also when I go in the Visa Test Panel in the USB Control (input/output) I get this in return data :
USB Control In
wIndex: 0x0
wValue: 0x0
bmRequestType: 0x80
bRequest: 0x0
wLength: 0x0
Error
VISA: (Hex 0xBFFF0078) La valeur de certains paramètres n'est pas valide (ces paramètres ne sont pas connus).
Is someone capable of helping me resolve this ?
Thank you.. 🙂
06-18-2015 11:55 AM
Hello,
A time out error is saying the scope did not respond to something. (Or, you have tried to read something from it when there is nothing there to read).
The error you are getting from the VISA tool is because you are not using it correctly.
Go to the "Basic I/O" tab and hit the "Query" button and you will see the scope respond to the "*IDN?" query that is there by default:
i.e. Not this:
Instead, use this:
Regards,
Leonard Brown
Applications Engineer
Teledyne LeCroy
1-800-553-2769
06-19-2015 05:58 AM - edited 06-19-2015 05:59 AM
Thank you ! The problem is solved ! 😄