The frequency of my signal is around 20kHz,but the frequency can vary by about 10 Khz depending on the crystal conditions.
02-22-2020 11:24 PM
Why do you multiply by 1M?
It seems like with a 80MHz clock, your count is number of 1/80M seconds in a full period.
As a result, to get Hz you need to take the reciprocal (as you do) then multiply by 80M, not 1M. Even if you want the result in kHz, you have the wrong multiplier (then you'd want 80k).
Otherwise, the precision of your measurement is determined by your SCTL click rate, which is already high enough to give good resolution at the frequencies you describe.
You might consider some averaging if your values are varying too quickly.
02-23-2020 02:50 AM
Hi,
you are right about the moltiplication by 1M, but is simply a mistake, I attached an old version.
Anyway, whit 80 Mhz of clock I' m able to appreciate the Hz variation till 4K, above this frequency the precision around the Hz begins to decrise.
Furthermore, just yesterday I have known that the frequency range has been moved from 3Khz to 100Khz, so now I have the same needs (1 HZ sensibility) but in a bigger frequency range.
I know that is it possible to use two counters to increase the sensibility in a frequency measurements, but I don't know how to implement it in FPGA.
Can you halp me?
regards
David
02-23-2020 03:01 AM
I'll take a look at your code when I return to a PC, but in case you haven't read it, you might find the following link helpful: Making Accurate Frequency Measurements
02-24-2020 04:06 AM
Hi and thanks for the link.
I would like to use the 2 Method(Count Number of Pulses in Known Time), I think should be the best solution for my pourpose,but I don't know how to implemet it in FPGA do you have any idea?
Regards
David
02-24-2020 09:41 AM - edited 02-24-2020 09:49 AM
Here's a simple example showing the logic on Desktop (so with a Wait, no SCTL, boolean control as input). There is limited debugging but you could add more indicators to display updating values if it isn't clear.
02-25-2020 07:42 AM
with two more shift registers you should get an even better reading 😉 and since the first edge is the last previous you could even extent the gatetime (10s? ) to get a fast and a higher resolution freq. counter
the false cases just pipe the values
02-25-2020 08:29 AM
Thanks for your input,I have done a new VI (See Fig 1 and Fig 2)strating from your sketch,but some things are not clear to me :
1) when yo said"choose a meseurment time" means choose a FPGA onboard clock?
2) What is teh input of first register(Measurement Ticks)?
3) Apparently if I choose 80Mhz as onboard clock, and 80Mhz as Masurement Ticks the "Number of edge " is dierctly the frequency is it correct?
Do you have some Advice to improve the VI?
thanks
David15
02-25-2020 09:25 AM
Hi Henrik,
Your solution is bit more complex for me, can you explaln it better ?
I tried to develop your VI but maybe I mada some mistake because it didn't work properly.
Please find attached the VI can you check it?
Regards
David
02-25-2020 10:07 AM
@David15 wrote:
Thanks for your input,I have done a new VI (See Fig 1 and Fig 2)strating from your sketch,but some things are not clear to me :
1) when yo said"choose a meseurment time" means choose a FPGA onboard clock?
By a measurement time, I meant a period of time (in seconds, for example) over which you want to make the measurement. This is a free choice, but has the following tradeoff: the longer the time, the more accurate (advantage) but the less frequent (disadvantage) - because you will get one result per measurement time.
@David15 wrote:
2) What is teh input of first register(Measurement Ticks)?
Measurement Ticks is the number of ticks that correspond to the measurement time. So if you have an 80MHz clock, and want measurements every 0.1s, then you'll have measurements at 10Hz rate and your "Measurement Ticks" value would be set to 80MHz * 0.1s = 8M.
@David15 wrote:
3) Apparently if I choose 80Mhz as onboard clock, and 80Mhz as Masurement Ticks the "Number of edge " is dierctly the frequency is it correct?
Yes - this is true. If you choose the same number of ticks as your clock rate, you'll get one measurement value per second. As a result, the number of edges per result will be the same as the number of edges/second, or frequency.
@David15 wrote:
Do you have some Advice to improve the VI?
I guess you mean the input being 100320Hz and the result being 100322 Hz? I think this almost lies within your error:
Fk = 80MHz, Fx = 100320Hz, error = Fx+-1Hz (since you're measuring for 1 second, the difference between adjacent counts is 1Hz). If you increased Measurement Ticks, you'd see this decrease, but you'll take longer to make a single measurement.
You can also try some variations (which perhaps include Henrik's post, I'm still working it out) to reduce the error further without reducing measurement frequency (for example, by checking the number of 80MHz edges in a known number of the slower ticks, rather than only using the total number of slower ticks in a specified number of faster ticks). It's not clear to me if this is method 3 in the link I gave previously - that description is actually a little unclear to me in general...
@David15 wrote:
thanks
David15
You're very welcome. Feel free to give kudos to helpful posts 😉