08-28-2012 04:22 AM
Hello, Ravi Antani,
I am currently using a E2V ethernet gigabit line scan camera (8 or 12bit resolution). I could get images by using it. But when I tried to get image raw data by using IMAXdx get image data.vi, I got the following results:
(1) camera set at 'mono8' resolution
I got the right image data size, i.e. 2048*100 when I set the width 2048 and the height 100. The 1D data is like 8, 8, 8, 8, 7, 8, 8 --------.
(2) camera set at 'mono12packed' resolution
I got the image data size: 3072*100 when I set the width 2048 and the height 100. The acquired 1D image raw data is like 8, 238, 8, 8, 238, 8, 9, 160 --------.
Could I ask you how can get the right image raw data 2048*100 from the readout data 3072*100?
Many thanks.
Zeng
08-28-2012 04:22 AM
Hi there Zeng,
So typically, since this is a relatively separate issue from the previous one, we like to have you create a new thread on the forums. That way, more people are likely to look at it.
I will let you know that this might depend on the specific camera you are using. The Get Image Data function always outputs 8-bit values no matter how many bits the pixels are. What's probably happening is that width-wise, alternating pixels will have 4 bits from the value before, and 4 bits from the value after.
For example, Value (0,0) in your array has the first eight bits of Pixel (0,0), and then Value (1,0) has the last four bits of Pixel (0,0) and then the first four bits of Pixel (1,0). Value (2,0) then has the last eight bits of Pixel (1,0).
I hope this makes sense. You might have to rearrange your data using a loop into a U16 (to fit all 12 bits).
08-28-2012 04:23 AM
Hi, Ravi,
Many thanks for your support, but I still did not work out this problem:
The actrual pixel value should be 8, 8, 8, 8, 7, 8, 8 -------- with camera resolution set at 'mono8' or 'mono12packed'. The value could be rewritten in bits at 8bit like:
Pixel(0,0): 0000 1000
Pixel(1,0): 0000 1000
Pixel(2,0): 0000 1000
Pixel(3,0): 0000 1000 --------
The value could be rewritten in bits at 12bit like:
Pixel(0,0): 0000 0000 1000
Pixel(1,0): 0000 0000 1000
Pixel(2,0): 0000 0000 1000
Pixel(3,0): 0000 0000 1000 --------
Now following your advice, the readout values when camera was set at 'mono12packed' are like:
Readout(0,0): 0000 1000 (the first eight bits of Pixel (0,0) at 12bit), the value is 8 (correct because in agreement with the readout value at 8bit);
Readout(1,0): 1000 0000 (the last four bits of Pixel (0,0) and then the first four bits of Pixel (1,0) at 12bit), the value is 128 (incorrect because the readout value is 238 at 'mono12packed' );
Readout(2,0): 0000 0000 (the last eight bits of Pixel (1,0) at 12bit), the value is 0 (incorrect because the readout value is 8 at 'mono12packed' )
Could you please give me more explanation on where is wrong in my analysis above?
Thanks. Zeng
08-29-2012 04:29 AM
Solved. Thanks.
