05-12-2009 03:42 PM
Hi all
I have encounter a problem like in the topic http://forums.ni.com/ni/board/message?board.id=170&message.id=404676&query.id=250408.
I'm sorry for repeating of topic but I've read all of your advices and have not yet resolved problem.
My task very simply: I use USB 6251, differential mode to measure voltage of battery Energizer 1.5V. But result shows >10V ??? Mean.vi shows bigger 4V (see the attached picture)
I don't understand what happed with my task?
Please help me!
Thank so much!
MC
Solved! Go to Solution.
05-13-2009 04:53 PM
Hello,
Thank you for posting to the NI Forums!
How do you have your battery connected? Are you using bias resistors? This could be a grounding problem and may be fixed if bias resistors are used. Check out Table 1 in this document and the differential configuration for a floating source.
05-13-2009 10:59 PM
Thank Flaugher
I did not connect bias resistors. But now I do another way: connect possitive pole of battery to the AI0 and other to the AI GRD, choose input configuration "RSE" and get good results.