Issue with noise on an analog signal

I require assistance on a LabVIEW issue I am experiencing and hope somebody can provide a solution.

I am trying to measure an analogue differential signal from an external device. I have 4 devices in total that I am measuring at the same time. Each device has its own separate USB-6001.

I am trying to measure the period and pulse duration of the signals. This should be a square wave pulse with period of 2 seconds and pulse duration of 50ms.

Due to noise on the line, LabVIEW is finding it difficult to measure these readings correctly.

Please see 2 attachments:

1. VI being used to measure the analogue signal.

2. Screenshot showing the outputs of 4 tested signals (measurements in question shown in green circles). The first signal is ok with minimal noise. However, the next 3 signals, which are measured in the same way are noisy.

 

Is there a way on LabVIEW of increasing the threshold on the measurements so it can be raised above the noisy section (say around 4 looking at the screenshot).

I have used a peak detector to measure frequency earlier in the code but I cant get this to work in the same way.

I have also tried to fit capacitors to filter out the noise but this has also made no difference.

I have also moved the 4 devices around and the clean reading is always on the top left in the first position.

Any help would be much appreciated.

 

Thanks. 

Download All
0 Kudos
Message 1 of 7
(1,278 Views)

Are you using the USB cable ferrites?  You have a ground loop problem! Hardware, not Software design is needed to correct this.


"Should be" isn't "Is" -Jay
Message 2 of 7
(1,249 Views)

Yes I have a ferrite cable installed between the DAQs and the PC. Do these need to be installed anywhere else? I believe the unwanted signal is coming from the device being tested, which is why I was asking if there was a 'workaround' in software to get a more accurate measurement. I also can't understand why the position 1 reading is a lot better than my others

0 Kudos
Message 3 of 7
(1,215 Views)

Try using a butterworth filter. See attached VI

Message 4 of 7
(1,193 Views)

Thank you for the suggestion. This has been implemented successfully. 

I have also ordered clip on ferrites to try and eliminate the noise completely on the test lines.

 

Thanks a lot for your help.

 

0 Kudos
Message 5 of 7
(1,147 Views)

These appear to be binary signals and you don't care about amplitude, so you can just threshold your data. For each datapoint, if it's less than 2.5V, call it a 0. If it's more, call it a 5. That should get rid of your noise problems (though the hardware filter will help too).

0 Kudos
Message 6 of 7
(1,125 Views)

That makes a lot of sense, thank you for your feedback and suggestion it is much appreciated.

0 Kudos
Message 7 of 7
(1,063 Views)