I have two analog output AO Series devices connected via RTSI.
I have defined the RTSI Cable in MAX and added the two boards.
I am having trouble getting the slave to use the master Sample Clock.
According to 'Timing and Synchronization Features of NI-DAQmx', synchronization will be seamless.
Attached is my sample VI. Device 1 finishes, but Dev2 never does.
Any ideas?
I am using Windows 7, Labview 12
thanks
Solved! Go to Solution.
10-08-2013 10:55 AM
Hi GretchenSkellyZimmer,
If you've added your RTSI cable to MAX and added the devices to the cable, the sample clocks should be shared as you've set your code up. Do both analog output tasks start and can you see output on it? Are you triggering your output to start?
Regards,
10-08-2013 11:15 AM
Thanks for your reply. Yes, I have added the cable to MAX (see image below - also tells you what AO Series device I have)
I am using 'No Trigger', so I assuming Dev1 is triggering, also going by the fact that the Dev1 Boolean goes green. Dev2 Booleen does not go green.
10-08-2013 01:48 PM - edited 10-08-2013 01:48 PM
If you save it in LV 2011 or post a screenshot I could take a look (the project I'm working on is still using 2011 so I haven't upgraded yet).
Best Regards,
10-08-2013 02:13 PM
Thanks for the offer!! Here you go,
Gretchen
Looks familiar, right? Looks like the DAQmx Examples.
Here's the snippet
10-08-2013 03:08 PM
Assuming the "Enabled" case of your diagram disable structure just passes through the task handle and error wire, the software looks good (which is actually unfortunate, I was hoping it would be something obvious). You're not getting any software errors, right (the screenshot seems to indicate that you are getting no errors)?
As a troubleshooting step, could you try polling the Total Samples per Channel Generated property for Dev 2 inside your loop?
If the samples generated number stays at 0, Dev2 isn't seeing the clock at all.
If the samples generated number increments to some number (0 < N < 1000) then Dev2 is seeing some of the clock edges but missed some. Try the same steps as #2 above (bandwidth issue).
If it increments to 1000 then Dev2 saw all of the clock edges but the task isn't reporting that it is done. If this is the case, somebody from NI would need to look into whether there is a bug with Is Task Done? when using an external clock on the 6723. I don't have the hardware myself to check. If this is your scenario, the sample clock was received properly and all samples would have been generated so it's just a matter of finding another way to exit your loop (could use Total Samples Per Channel Generated for example).
If it increments to >1000... well this just wouldn't make any sense since Dev1 is only supposed to issue 1000 sample clock pulses. There would have to be noise on the RTSI line for this to happen in conjunction with a bug in DAQmx causing the Is Task Done? to not work with an external clock on your device.
Best Regards,
10-08-2013 03:21 PM
THE ANSWER IS ..........
TotalSampPerChannelGenerated = 999 (I asked for 1000.)
Yikes! What does that tell you?
10-08-2013 03:56 PM - edited 10-08-2013 03:58 PM
If you're consistently getting 999 samples then it's not a bandwidth issue (so ignore half of my previous post).
There's some bug or undocumented feature of the 6723 that's giving you this behavior. The device is pretty old and I'm not too familiar with its inner workings, but I'll take a wild guess and speculate that your last sample is sitting in the latch and requires an additional sample clock edge to push it through (if this is true, it would mean that the synchronization was actually off by a whole sample clock period!):
Instead of sharing a sample clock, I would configure your tasks like this:
This way each device will generate its own clock and do what it needs to output all of the samples. The tasks will start synchronously due to the shared start trigger and will not drift over time due to the shared timebase.
Best Regards,
10-09-2013 01:06 PM
Thank you SO MUCH, John_P1!
I implemented your suggestion (using shared timebase), and it worked!
Dev2 completed right away with 1000, not 999. Should test with scope to make sure sine waves are lined up.
I then made another test: I asked Dev1 to wait for a trigger. both devices waited, and I used an external trigger to Dev1 and both devices responded.