One external pulse on PFI 0 to generate one sample point on an analog channel independent of the pulse rate (4Khz max)

Solved!
Go to solution

I haven't had much luck finding a suitable labview vi replacement for what I was doing with data acquisition using Component works CWAI acquire in VB. I have a tube line with variable speed and wish to sample at a fixed distance along the tube. Currently I sample every 1 mm of tube length independent of the linespeed (Capstan driven). The examples I have tried that utilize PFI0 as the trigger and/or clock are still asking for a sample rate and then recording the data to TDMS or Excel in units of the input sample rate not a variable rate which would come from a variable rate PFI0 if it was truly controlling sample rate. Perhaps someone out there can point me to a better example. (My PFI0 is sourced from an encoder pulse attached to a variable speed capstan. I am testing with X series USB-6356. Successfully been using PCIe 6023-6025 for last 20 yrs on production line with CWAI acquire)

 

Thanks.

0 Kudos
Message 1 of 3
(2,502 Views)

While I haven't used PFI 0 to generate a single sample point, I have used it to trigger DAQmx to collect a set of N Samples from M Channels at some fixed clock rate (i.e. 100 samples at 1KHz from 4 channels).  "Reasoning by analogy" (a sure-fire way to screw up!), I see no reason it couldn't trigger 1 sample from M channels.  I was using a USB-6009 or USB-6002, a fairly simple DAQ device.

 

What I'm less certain about (without doing some tests, and who has time?) is what the frequency limits on the PFI 0 pulses would be.  Should be simple to write a tiny program and test it, I would think ...

 

Bob Schor

0 Kudos
Message 2 of 3
(2,460 Views)
Solution
Accepted by topic author bobInohio

1. It should be fine to use PFI0 or PFI-whatever as an external sample clock signal.  You'll get one sample per active edge.  With the clocking signal coming from an encoder, you get data that's equally-spaced in *position* rather than in *time*.

 

2. You need to step in and take control over what's being written to file.  Wherever you see a waveform dt or a time vector being generated, step in and override it with your own dx of 1.0 mm.  Then train yourself to think of the data as it relates to *position* rather than the typical *time* seen in most examples.

 

3. [Advanced].  If necessary, you'd also be able to capture timestamps for those encoder pulses with a counter task.  It takes a little learning curve to get the tasks configured and sync'ed correctly, but it's very possible if it turns out you also need timing or speed information.

 

 

-Kevin P