05-01-2023 08:06 AM
Hello,
I have a PXI-4065 DMM with a PXIe-1088 chasis. After running the unit I can see it updates every second. I went and check the specs and it says if I have a resolution of 6-1/2 the reading rate is 1S/s which makes sense. But I need it to be faster than that so I tried changing the resolution to 4-1/2 which says the reading ratee should change to 100S/s. Unfortunately the rate is not getting faster than every second. How can I adjust that rate??
Solved! Go to Solution.
05-01-2023 08:42 AM
In almost all cases, it is because you've configured things not the right way. It would be helpful to look at your code to see where you are missing.
Please attach your VI in LV2019 or older.
05-01-2023 09:17 AM
thanks for your help. I found a YouTube video that mentioned about modifying the aperture time and the settle time and it worked. I was missing that part.