05-14-2009 12:46 AM
Hello all,
I tried to controll a keithley 2400 source meter with Labtracer using a NI GPIB-USB-HI card but when I run the test is send me this error message: "There was ab error reported on Visa Read in Query. Test aborted. Error code: -1073807339"
is this card (NI GPIB-USB-HI) able to controll keithley 2400 source meter ?
the keithley 2400 source meter screen show "REM TALK" during the running of the program,
I installed NI488227 in my computer (windows XP) and also "Agilent Connection Expert", this later also show no possibilty to connect with the Keithley meter, I attached a file for "Agilent Connection Expert".
Any help is appreciated
Thank you by advance
05-14-2009 03:34 AM
Follow the preliminary steps to first check if the device is able to communicate via the GPIB-USB convertor.
navigate to: Start>>National Instruments>>NI 448.2>>Interactive control. Use this utility to query the GPIB device for some standard commands for e.g.: *IDN?
There is also a GPIB analyzer in the above path which you can use to verify the device connectivity to the system.
Also for communicating the with 3rd party GPIB devices always enable " NIVisa Tulip.dll" which will help recognising the device in max. Look at the picture to enable this option in max.
post back for queries
05-14-2009 09:10 AM
Installing the Agilent software was not at all necessary and may be the cause of your problem. The Agilent software would be used if you had an Agilent GPIB controller. You may have installed some software that replaced something required for an NI controller. Why did you think you needed the Agilent software in the first place? I would suggest you remove everything related to Agilent and reinstall all of the NI software.
JK,
I'm going to have to disagree with you about tulip. It's only necessary when you want to use a third-party GPIB controller with MAX and LabVIEW. It does not affect communication with instruments.
05-15-2009 07:16 AM
JK and Dennis,
Thank you for your answers,
I tried to communicate with the meter by the method suggested by JK but I failed, when I used the GPIB analyser it send me this message:"A GPIB Analyser interface was not found, you still may use this application to view previously captured data. Refer to the NI-488.2 Readme file for a list of supported analyser interfaces for your operating system" and the window for GPIB Analyser was opened with this title: "Unknown/Missing GPIB+card""
I removed everything related to Agilent and reinstalled NI488227 again but still the problem persist,
Thank you
05-15-2009 09:22 AM
You can only use the analyzer if you have a '+' type card that includes the analyzer functionality.
At this point, I really don't know where and how you are getting an error. Your very first step is to use MAX to verify that the driver is installed correctly and the controller is recognized. These are the basic steps in your installation instructions. Then, once the controller is recognized and passes those tests. Do not do anything until this works correctly. If it does work, right click on the controller and select 'Scan for Instruments'. If that fails to find your instrument, post back.
05-15-2009 11:01 AM
Yes it fails to find the instrument, attached is a print screen,
thank you for your help
05-15-2009 11:16 AM
05-15-2009 11:40 AM
05-15-2009 12:04 PM
This time I tried again with the Keithley 2400 source meter, it was detected (I don't know what was changed with before) but it didn't respond to a query, and the same problem like before remain for Labtracer (same error message)
05-15-2009 12:57 PM