Did you choose a valid physical channel, cjc channel and cjc source as internal??
Plus there is one more example vi called SCC-TC0x.vi ( uses Traditional DAQ device drivers), try acquiring with this
( Note; you may have to reset Device if you have to use DAQmx functions after using traditional DAQ functions)
Also, grounding might be an issue. So, try connecting the ground of SC-TC02 to the SC-2345 ground.
09-20-2006 04:44 AM
09-20-2006 05:20 AM
Have you installed traditional (Legacy) NI-DAQ device drivers??
If you have, you will find this attached example vi
Note: it will work only if you have installed traditional DAQ device drivers.
09-25-2006 07:33 AM
Sorry for the late answer...
I don't have thoose drivers installed so it doesn't work.
But I was thinking, can't I use a normal voltage input instead and scale the value to represent a temp. based on K-thermocouple?
09-25-2006 07:54 AM
You mean without a cjc??
that will not give you a valid/ correct thermocouple measurement