08-13-2010 08:13 AM
I tried the same and getting timeout error.
I couldn't find any help for now in TEK site also.
Is there any update or solution for this issue.Please find the attached image of NI SPY.
I tried with some delay of 10 seconds after starting the hard-copy.
The problem is that i don't get the image from the GPIB using the "Read to File" function except for the timeout.
Any suggestion or help is greatly appreciated. Thanks in Advance.
Thanks,
Davidson
08-16-2010 05:50 PM
Hi Davidson D,
What version of LabVIEW are you using and what version of the Tek 3054B are you using?
08-16-2010 10:11 PM
Dear Joe,
I use LabVIEW 8.6.1f version.
And TEKTRONIX,TDS 3054B,0,CF:91.1CT FV:v3.26 TDS3GV:v1.00 TDS3FFT:v1.00 TDS3TRG:v1.00
Thanks,
Davidson
08-16-2010 10:13 PM
Hi Davidson,
Were did you get the instrument driver?
08-16-2010 10:35 PM
I hope the driver was downloaded from NI site. Its "tkds30xx.llb".
Actually this driver didn't have the screenshot options. So i was using the commands in manual "TDS30xx User Manual.pdf" and created a VI for the hardcopy function by sending the commands i highlighted in the SPY.
Thanks,
Davidson
08-16-2010 10:48 PM
Instead of doing a Read To File, try doing a normal VISA Read and set the VISA Timeout to a higher value.
You could also just save the LabVIEW image of your acquisition. That way, you do not transfer the same data twice.
08-16-2010 10:50 PM
Dear Joe,
Also i installed the latest driver "Tektronix TDS 3000 Series" - Project style from IDNET. It do not have Hardcopy functionality in it 😞
Thanks,
Davidson
08-16-2010 10:56 PM
@Dennis,
I will try this with the instrument and update what happens with the normal VISA read.
We are using the "save LabVIEW image" method temporarily.
Thanks,
Davidson
08-18-2010 10:59 PM
@Dennis,
VISA Read is also getting timedout.
Does anyone have the working version of this VI for screenshot?
Regards,
Davidson
08-23-2010 03:38 AM
I got it working.
It was my mistake of sending the HardCopy start command incorrect.
I was sending HARDCopy:STARt instead of HARDCopy START
I should have a lens to verify the commands that are sent to the scope.
Thanks,