Hello,
I am using an USB-8451 device with LabVIEW 8.6. I have installed the 1.1.4 drivers of such device and both windows and MAX detect it. I have also renamed the alias and the status of the device is "This device is working properly".
The problem comes when I try to add a DAQ-assistant block to my project. I select the kind of input I want to have (digital line, though I tried them all) and LabVIEW does not detect any device.
I tried to do a diagnostic with VISA Test Panel. There I guess something is wrong, as when I execute a viRead i get as Return Status 'BFFF0015h' in red letters... but I do not really know what does that mean. The LED on the device is green and blinking.
Thank you in advance.
Solved! Go to Solution.
03-22-2011 08:05 AM
Can you address it directly by using a channel constant?
VI chain : Create a task, config channel*, run task, read from port?
03-22-2011 11:49 AM
It's not supposed to show up in the DAQ Assistant (and NI-VISA is certainly wrong) or haven't you read the manual?
03-22-2011 09:07 PM
You will have to write your own Labview code using the drivers that come with the USB-8451. There are no Express VI's built for that instrument. It is not very complicated if you read the manual as Dennis has suggested.
03-23-2011 02:52 AM
Thank you all for your replies... I had the idea of having used this device through DAQ-assistant in the past, so I did not go through that part of the user guide. Dumb of me.
Any idea of why can be the VISA analysis wrong?