cannot find network device

i have tried to connect the device using MAX. I know the IP address for this ethernet device but not the port #. i used '7' as given by default. it connects with but with yellow mark on it. In MAX, when i right click on this TCPIP device there is no 'communicate with device' option. when I click 'open visa test panel' following is the message:

 

Unable to open session to "TCPIP::192.168.1.104::7::SOCKET"

Return status code: 0xBFFF0011

Status name: VI_ERROR_RSRC_NFOUND

 

Insuffcient location information or the requested device or resource is not present in the system. 

 

does this mean that this bruker device can only talk with the software provided by company, not anything else? 

0 Kudos
Message 11 of 13
(658 Views)
No, it means that the bruker is expecting it's communication on a specific port. You need to know what that port number is. You can't simply pick one out of the air and expect to connect to a server (which the bruker is since you clients connect to it).


Mark Yedinak
Certified LabVIEW Architect
LabVIEW Champion

"Does anyone know where the love of God goes when the waves turn the minutes to hours?"
Wreck of the Edmund Fitzgerald - Gordon Lightfoot
Message 12 of 13
(654 Views)
Thanks. I could finally find the port number and now MAX connects to this field controller successfully. Thanks to all the contributors.