daq assistant calibration drift

Helloo,

 

I am using the USB 6229 DAQ. In my VI I am using the DAQ assistant to calibrate my device before I begin recording. However I am finding that there is a constant drift in the 'uncalibrated value' before I commit the new values that I want to use.

 

Is this drift normal or is there a problem with my DAQ?

 

any help would be appreciated,

 

Al

0 Kudos
Message 1 of 4
(2,378 Views)

How much drift are you measuring?  Generally NI devices seem to be quite stable.  Are you allowing sufficient warm up time?

 

Lynn

0 Kudos
Message 2 of 4
(2,371 Views)

Hi,

 

The uncalibrated figure is approximately '0.4523' and it constantly reduces in size. If i leave it for 5 mins it will be around 0.3800. It doesn't seem like a majoy drift but the uncalibrated max and min points of my device are around 0.2 and 0.6.

 

The DAQ had been on for about 7 hours and i was detecting a drift so I restared everything and tried again numberous times over 2hrs and i detected the same drift in calibration.

 

any ideas?

 

Al

0 Kudos
Message 3 of 4
(2,366 Views)

What is connected to the input while this drift is occurring? What is the input configuration, differential or single ended?  What is the full scale range?

 

Lynn

0 Kudos
Message 4 of 4