Just put the PXI card into appropriate chassis, connect it with appropriate cable with the BNC adapter and the output from your bridge module connect to one of the channels on the BNC adapter. The only question is your bridge realization but I suppose you use some commercial solution like this that gives you normalized voltage output. In case you built or you are going to build your own Wheatstone bridge realization you will need to amplify the output from the bridge because it gives voltage around mV. Than you just simply acquire the voltage. Some of the examples shipped with LabVIEW can be handy. Then of course you have to scale and compute the data according to what you are trying to measure (deformation, strain, force...)
Hope it helped
03-10-2008 01:39 PM
Hi pjoule,
Not only will you need to amplify the output, you need excitation for the Wheatstone bridge. Ni.com/strain is a good starting place to learn about what is needed for strain measurements (“Strain Basics” in the upper left in blue) and also what hardware can be used. Not only can SCXI be used, but there are some newer technologies such as CompactDAQ (more portable) and PXI (newer than SCXI). These can be found by clicking on the “Choose the Right DAQ System” link, also from ni.com/strain.
03-10-2008 02:25 PM
03-11-2008 10:33 PM - edited 03-11-2008 10:37 PM
03-12-2008 08:24 AM
03-12-2008 11:39 AM
Hi pjoule,
I’m not sure if I
entirely understand the question. It sounds like you’re asking about providing
an excitation regulator, is that correct?
With our CompactDAQ (cDAQ) strain gauge module, the NI 9237, the excitation voltage is also the same voltage used for a reference to the ADC, so no matter how much the voltage fluctuates you will always get a consistent mV/V reading.
With you particular
setup, you would need something to regulate the excitation voltage, because the
DAQ card only reads voltage, not mV/V like the NI 9237, so if your excitation
is higher or lower, you reading would also be higher or lower. Again, what you
have isn’t specifically designed for strain gauge measurements, so you’ll need
to be careful how you do your measurements.
I hope that helps.