labview 5.1 and ni-daq 6.9.1 acquisition problem

Hi all,

 

I have a DAQ acquisition system with the two software versions above specified. I am trying to acquire a sinusoidal wave of 240Hz and 5V amplitude.

 

These are the steps I have followed as advised in the manual.

 

1) Function generator outputs a sine wave of 240Hz and 5V pk-pk

2) I have wired the signal to the breakout board for a differential measurement on channel 0: + on ACH0 (pin68) and - on ACH9 (pin34)

3) As acquisition parameters I set sampling rate = 200*240 = 48000Hz and data points 200*4 = 800 in order to capture 4 whole periods with 200 data points each

4) I set Device 1 and Channel 0

 

From labview I can just see a constant line at 10v then using the test channel from ni-daq software I can see that the input to the acquisition card is a step value which jumps between 0v and 10v, but no sinusoidal behaviour at all.

 

The labview project VI works correctly with other acquisitions (temperature measurements) so I can say it's not the problem.

 

I don't really know what other parameter I could change. Anyone has an idea of what is the problem or how I can diagnose if the acquisition card/breakout board has any hardware issue?

 

Thanks in advance!

0 Kudos
Message 1 of 18
(3,448 Views)
Is that a typo or do you actually have AI0 and AI9 connected? For differential, it should be AI8 instead of AI9.
Message 2 of 18
(3,433 Views)

Hi,

 

yes sorry it was a typo.. pin 34 is ACH8 (negative for differential on channel 0).

Any clue why I get a straight line at 10v in my acquisition screen? From a DSO hooked up on the same signal I can clearly see the sine wave at 240Hz and 5v pk-pk.

0 Kudos
Message 3 of 18
(3,419 Views)
You can use MAX to test things without programming. But I will guess your configuration is wrong. Check out the Daq card user manual for recommend ways to connect to your system with different input configurations


Besides which, my opinion is that Express VIs Carthage must be destroyed deleted
(Sorry no Labview "brag list" so far)
0 Kudos
Message 4 of 18
(3,413 Views)
By the way, you can use newer versions of Labview together with ni-daq 6.9.3. I have used Labview 8.6 and version 6.9.3 without any problem


Besides which, my opinion is that Express VIs Carthage must be destroyed deleted
(Sorry no Labview "brag list" so far)
0 Kudos
Message 5 of 18
(3,408 Views)
HI,

WHERE I CAN FIND THIS MAX? ALSO I FOLLOWED THE CONNECTION PROPERLY WHICH IS PRETTY STRAIGHTFORWARD, ONLY TWO WIRES, SO I NEED TO TEST THE CARD

IS THERE ANY FREE UPGRADE OF LABVIEW?
0 Kudos
Message 6 of 18
(3,398 Views)

You can look for the file here .........program files\National Instruments\MAX\NIMax.exe

And yes Labview has been updated a lot since Labview 5.x 😉 But those updates are nor free

Message Edited by Coq Rouge on 04-15-2010 05:05 PM


Besides which, my opinion is that Express VIs Carthage must be destroyed deleted
(Sorry no Labview "brag list" so far)
0 Kudos
Message 7 of 18
(3,389 Views)

MAX is Measurement & Automation Explorer and should be on your desktop assuming that the version of the DAQ driver is not so old that MAX was not used. I don't remember that far back.

 

No, there is no free upgrade of LabVIEW.

0 Kudos
Message 8 of 18
(3,388 Views)

Thanks for your help, this morning I got it worked but I still think the card has a problem. The connections are correct since when it works I can see the right sine wave.

 

As I said I believe there is some sort of hardware issue in my system: I went to Measurement & Automation Explorer, under device, I tested the device. I got an error saying that the primary interrupt has a problem. I went ahead and in the windows of the device test panel I could see the signal correctly. I started labview and took my measurements. After 5 min of acquisitions, in the windows of the test device of Measurement & Automation Explorer the signal went back to a straight line at 10v and of course I couldn't read anything anymore in labview project (only the straight line at 10v).

 

I closed all the programs, wait 5 minutes and restarted Measurement & Automation Explorer, tested the device again, aquired with labview and everything was fine... Till again a straight line after some time.

 

I guess there is a problem of suspension/hardware on my system. I will try MAX right now, but is it normal I have to costantly test my device to push a reading?? Indeed if I don't follow this procedure, labview cannot read anything.

 

Thanks!

0 Kudos
Message 9 of 18
(3,373 Views)

So I figure out through MAX, i.e. the explorer, that my card sometime read a constant value of 10v, while other times it correctly see the wave form.

 

When I start testing my device (PCI-6023) from MAX, I got this error:

 

"The device is not responding to the first IRQ level. Continue with the test panel?"

 

Saying yes to this message brought up the panel but again in a total random way the card sometime 'see' the waveform other times 'see' constant value of 10v...

 

Any idea about that?

 

Also I have another machine with the latest version of labview, but I don't think I can open my VI project or that this is compatible without recompiling. Do you have any suggestion about that either?

 

Thanks!

0 Kudos
Message 10 of 18
(3,363 Views)