seebeck coefficient Measurement
T1 and T2 or Thot and Tcold is read by Lakshore 336 temperature controller. and dV from K182 voltmeter. as you know I also control T1 and t2 temperatures from lakeshore, I mean temperature graient. What do you think about it?
Message 31 of 32
(768 Views)
Hi
If you connect the dV wires to the entry of the lakeshore, you need to be sure that the two connectors at at the same temperature. I suppose the connection is made with copper to the nanovoltmeter.
You measure dV through the + wires. Suppose that initially you have T1=100 °C and T2=100 °C.
Increase a bit T1,
you get T1=101 and T2=100, dT=1K and you measure dV1
then T1=102 T2=100, dT=2K and measure dV2
and so on (I am applying a sinus gradiaent, and get something like 50 to 100 points, with a max dT of 3 or 4)
then plot dV versus dT, get the slope value m
the intercept should be zero but it will be not (offsets, that s why you need to measure the slope)
You get the slope : to this value you add (wire correction) for the contribution of the + wire
Look in the graphs I sent you before, at 100 °C the contribution of the + wire is 28 microvolt
m+28 microV is the value of the Seebeck coefficient of your sample.
Try with a sample (Pt wire, or Ni...) and post back if you need !
regards
Message 32 of 32
(765 Views)