01-16-2012 03:41 PM
01-16-2012 03:59 PM
Possibly this is due to the thermocouple wires picking up signal from the heater wires, although if the effect is a *sudden* transition at 800 C, this is indeed strange. Are the thermocouple wires twisted pair?
01-16-2012 04:14 PM
01-16-2012 05:14 PM
Do you have a graph of this phenomina? the loss of control at 800C is very close the the currie temperature of iron are you sure that is not a J type TC?
01-16-2012 05:20 PM
Twisting the wires around each other greatly reduces noise pickup from electromagnetic fields.
http://en.wikipedia.org/wiki/Twisted_pair
01-16-2012 05:37 PM
01-16-2012 05:42 PM
01-16-2012 10:29 PM - edited 01-16-2012 10:29 PM
Please use an oscilloscope than a multimeter to monitor noise, you wont be able to see noise with a multimeter!
01-17-2012 09:17 AM
Well there are a few other factors to consider then.
01-17-2012 05:35 PM
Hello, hope you are doing great!
Some simple things to check:
Be sure that the voltage range you are using for the thermocouple is the correct for the temperature rating.
You can use bias resistor to ground, or a simple cable to ground for each of the thermocouples
If you are using any connector block with gain setting be sure this do not make the measurement off scale.
Try different thremocuples.
The twisted pair seems like a good idea, however is there a chance you can measure the voltage the thermocouple is generating at 800C, it is a bit dangerous.
We can actually create a voltage task instead of a temperature task and check if we have the same behavior reading voltgae directly, If this is not the case we can create a custom scale for the linearity of the thermocouple and obtain temperature.
Luis A.
Applications Engineer.