10-23-2007 01:48 PM
System info: LabVIEW version 8.0, NI-IMAQ for 1394 2.0, Camera = Sony XCD-V50 (IEEE 1394b), Windows XP pro (service pack 2)
Working in Measurement and Automation Explorer, I encountered this:
"Sony 1394 XCD-V50" appears under Devices and Interfaces, but I get this error.
IMAQ error 0xBFF68001
"Cannot load the driver"
I've tried to configure the device by right clicking and selecting the driver "NI-IMAQ IEEE 1394 IIDC Digital Camera". I get this error.
IMAQ error 0xBFF68016
"Insufficient resources available for the required video mode"
I installed the Prosilica "Firewire Fix" utility to re-install service pack 1 firewire drivers to run 1394b at full speed. But the error remains.
10-24-2007 10:34 AM
03-28-2008 04:25 AM
03-31-2008 09:56 AM
04-29-2008 06:41 AM
04-30-2008 09:50 AM
06-18-2012 03:17 PM
I had the same problem. Tried to load drives got and error, can not update drive. If the camera show up under NI-IMAQdx devices then I may have the solution.
Click on the device, should be Sony XCD-V50, a screen will open. (if you see XCD-V50cr and you have a V50 wrong driver install. V50cr is for color camera)
At the bottom you should see some tabs
Click Acquisition Attributes
In Video mode, select 640 x 480 mono 8 15.00 fps
Click Grab, you screen may turn white, that mean camera is out of focus
The light below the firewire port on the camera should be flashing between red and green
this means it is recording.
Under the screen your frame number be increasing. Camera is working.
Hope this helps.