I am using a scxi 1102 module with the 1303 connector block. I have all 10 Mohm bias and pull up resistors in place. I'm measuring small floating voltages and they are only accurate when I connect the negative terminal input to chassis ground. Do I have to connect all 32 negative terminals to chassis ground or is there a single pin that will connect all of them to chassis ground. Thanks.
12-28-2011 10:07 AM
Hello Jon,
The SCXI 1102 is designed to read thermocouple measurements, but it can also read analog inputs. Since you're working with a floating source, have you tried changing the resistor networks to use the 10 ohm version as a pull-down? This will be much closer to connecting the negative terminals directly to ground. The manual (http://www.ni.com/pdf/manuals/321923b.pdf) says how to change the resistor network under "Changing Resistor Networks".
12-28-2011 10:11 AM
Colden,
Thanks for your response.
No I haven't tried installing the 10 ohm resistors yet. I need to find or order new ones. Once I get them I'll give it a try and report back.
Thanks again.