Multifunction DAQ
Bad differential channel on NI-6220
I have a NI-6220 card that appears to have one bad channel. When a +/- 2V sinusoid is run into it with a function generator the signal slowly drifts to -10V. The other 7 differential channels work just fine. Any thoughts?
Nick
Message 1 of 9
(3,520 Views)
How are you interfacing with the 6220? Using a SCB-68? A BNC-2120? A custom fixture?
Which channel is it?
The 6220 has one ADC that it uses in multiplexed mode; if all of the other channels are working then the ADC does not seem to be at fault, but one of the connections to it. Perhaps you have a bent pin somewhere that's not making contact properly, or perhaps your cable has a faulty wire?
Brandon Streiff
ni.com/compactdaq · ni.com/daq
Message 2 of 9
(3,508 Views)
The faulty channel is channel 1. This is using 0-7 numbering. We actually have the same issue on the same model card in a different machine. I will check the cable and the pins.
Message 3 of 9
(3,504 Views)
I did not properly enter the product number. It is actually a NI-6221.
The cable and the connections prior to the card are working properly. So, unless it is one of the contacts for the card/cable interface It is the card. I presume that the 6220 and 6221 are different so thank you for your patience.
-Nick
Message 4 of 9
(3,483 Views)
Nick,
What version of the 6221 do you have? USB, PCI, PXI, USB-BNC, OEM? And what kind of breakout module are you using?
Kyle Mozdzyn
Applications Engineering
Kyle M.
Message 5 of 9
(3,468 Views)
Kyle,
PCI
The breakout module is a CB-68LPR
The cable used is a SHC68-68-EPM
-Nick
Message 6 of 9
(3,462 Views)
Nick,
Sounds like youre either having a connection problem on the breakout module (be sure to recheck your connections, screw terminals are sometimes finnicky), or simply have a faulty channel, which may be need to be RMA'd if its keeping you from operating. If its the latter, please give us a call and we can go throguh the RMA process if you like.
Kyle Mozdzyn
Kyle M.
Message 7 of 9
(3,440 Views)
Kyle,
We are only using 2 channels so one faulty channel is not a problem as long as the integrity of the rest of the channels is not affected. Also, could you explain RMA.
-Nick
Message 8 of 9
(3,438 Views)
Accepted by topic author ndotson
Nick,
Good to hear that this issue isn't going to affect your application.
There is some good information on the Return process (RMA) here:
http://www.ni.com/services/warranty_repair_policies.htm
and here:
http://www.ni.com/services/warranty.htm
Kyle M.
Message 9 of 9
(3,435 Views)