Hi We are currently
in the process of upgrading our lab data logging system
from an old National Instruments system to a new CDAQ system.
In our work we do a lot of measuring of temperatures in electrical appliances…
in some cases we need to measure temperatures adjacent to mains voltage.
In our old system we used a programmable transmitter as protection of the data
logger from any accidental connection to mains supply.
I have attached a diagram of how the transmitter was connected to the old unit
and also the legend for the transmitter terminals.
Could someone review both and advise if this can be integrated into our new
system.
05-01-2013 10:13 AM
You could use your existing 5114 module for isolation and select a CDAQ current input module.
http://sine.ni.com/np/app/main/p/bot/no/ap/daq/lang/en/pg/1/sn/n17:daq,n24:cDAQ,n36:18/
-AK2DM
05-02-2013 04:44 AM
Thanks for that.
Will have to talk nicely to the boss and get him to spend some more money