Hi Gustavo,
Load cell measurements usually require some signal conditioning (bridge excitation,signal amplification, calibration,...) The USB-9215 doesn't provide this kind of functionality directly. You need some external HW which is doing that. If you want to use the 9215, then you must do all the signal conditioning yourself. NI doesn't provide load cell signal conditioning for the USB-9215. But there are other NI HW which can do that. Check this for more informations:
Load cell measurementsThe sampling rate of the USB-9215 is determined by the highest channel number read, not by the number of channels read.
Highest Channel Number Max. Sample rate
0 20 kS/s
1 10 kS/s
2 6.67 kS/s
3 5 kS/s
For example, if you scan only channel 3, the fastest you can acquire without receiving an error is 5 kS/s.
I agree with you that the documentation is not clearly enough and I hope that my answers can help you.
Regards,
Luca