03-16-2018 08:21 AM
Hi there,
If I have cDAQ 9179, full with 14 NI 9220 C Series Analog Input Module, would it be possible to acquire continuously the full 448 channels (32*14) at a sample rate of 100kS/s ?
If no, what are the limitations ?
Thx,
Solved! Go to Solution.
03-19-2018 04:54 PM
Hi,
Any answer ?
Thx
03-20-2018 03:29 AM
Hello Sébastien,
Have you seen that document?
What Is the Maximum Throughput On My cDAQ Device?
Does it answer your question?
Thanks
Junior
03-20-2018 09:13 AM
I think the last sentence in that linked pages is in error:
Because the total rate of data transfer during this application (8.6 MB/s) does not exceed the maximum throughput of the 9179's USB 3.0 bus (250 MB/s), there should be bandwidth issues acquiring data at these rates on this device.
It is missing 'not':
Because the total rate of data transfer during this application (8.6 MB/s) does not exceed the maximum throughput of the 9179's USB 3.0 bus (250 MB/s), there should NOT be bandwidth issues acquiring data at these rates on this device.
-AK2DM
03-20-2018 09:29 AM
You are completely right!
03-20-2018 10:21 AM
Yes, thank you very much.