Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

NI 9205 in a task; error -200077

Solved!
Go to solution

I'm trying to set up a task for reading all 16 differential channels on a NI 9205.

I get an error -200077 when I get to adding channel8 (starting counting on zero) to the task, almost as if the software

thinks this is a moduel with 16 single-ended and 8 differential channels, not 32/16 as in reality.

Anybody, anything ?

0 Kudos
Message 1 of 2
(3,893 Views)
Solution
Accepted by topic author MartinP

use AI0-7 and AI 15-23

 

AI8 is the - side of AI0


"Should be" isn't "Is" -Jay
Message 2 of 2
(3,890 Views)