08-02-2013 10:31 AM
Hi,
Here's my setup:
-developing code using LabVIEW 2012
-cDAQ-7142 with NI-9211 (calibrated) TC module using J-type TC
-measuring temperature range from 0-100 deg C
I am developing an ATE for a controller my company is developing. The product has a heater output and a TC input (J-type) to read the temperature of a material (in my case, a block of aluminum) that the heater is affecting. The ATE is my reference sensor to prove the temp. accuracy of the electronics in our product. For some reason, my NI-9211 is reporting a temperature reading that is > +/- 2 deg C that the IC is reporting back. The 9211 is calibrated & the ambient temp range is within the bounds that I don't need autozero enabled. I've searched the forums about CJC references and any other settings I might apply to both the module and my VI but still have a big gap in temperature readings.
I've used a FLUKE 87V with a K-Type TC connected and is reporting temp. readings within a +/-1 deg C. accuracy from the IC on the controller.
Any suggestions?
08-02-2013 12:10 PM
Type K thermocouple table, printed by Omega from data published by NIST:
"LIMITS OF ERROR
(whichever is greater)
Standard: 2.2°C or 0.75% Above 0°C
2.2°C or 2.0% Below 0°C
Special: 1.1°C or 0.4%"
I do not have the Type J table handy but I think the limits of error are similar.
Getting temperature measurements with less than +/-1 degree error with thermocouples requires considerable care and definitely requires system level calibration. Not easy.
Lynn
08-05-2013 11:34 AM - edited 08-05-2013 11:34 AM
It might be good to consider using an RTD for your measurements rather than a thermocouple. This document explains the difference between the two:
http://www.ni.com/white-paper/3778/en/
Jeremy P.
08-05-2013 01:31 PM
Unfortunately, I need to use a thermocouple for my application since I am verifying my controller's TC read back accuracy (which the IC has a tolerance of +/- 2 deg C) with my ATE. I'm still researching my solution and will (hopefully) post it here when I get one. Thanks all again.
Cristhian
08-05-2013 02:56 PM
While the controller needs a thermocouple, you could use an RTD for the tester.
Lynn