Here's my setup:
-developing code using LabVIEW 2012
-cDAQ-7142 with NI-9211 (calibrated) TC module using J-type TC
-measuring temperature range from 0-100 deg C
I am developing an ATE for a controller my company is developing. The product has a heater output and a TC input (J-type) to read the temperature of a material (in my case, a block of aluminum) that the heater is affecting. The ATE is my reference sensor to prove the temp. accuracy of the electronics in our product. For some reason, my NI-9211 is reporting a temperature reading that is > +/- 2 deg C that the IC is reporting back. The 9211 is calibrated & the ambient temp range is within the bounds that I don't need autozero enabled. I've searched the forums about CJC references and any other settings I might apply to both the module and my VI but still have a big gap in temperature readings.
I've used a FLUKE 87V with a K-Type TC connected and is reporting temp. readings within a +/-1 deg C. accuracy from the IC on the controller.
Any suggestions?
(3,620 Views)
Type K thermocouple table, printed by Omega from data published by NIST:
"LIMITS OF ERROR
(whichever is greater)
Standard: 2.2°C or 0.75% Above 0°C
2.2°C or 2.0% Below 0°C
Special: 1.1°C or 0.4%"
I do not have the Type J table handy but I think the limits of error are similar.
Getting temperature measurements with less than +/-1 degree error with thermocouples requires considerable care and definitely requires system level calibration.  Not easy.
Lynn
(3,613 Views)
It might be good to consider using an RTD for your measurements rather than a thermocouple. This document explains the difference between the two:
http://www.ni.com/white-paper/3778/en/
Jeremy P.
(3,584 Views)
Unfortunately, I need to use a thermocouple for my application since I am verifying my controller's TC read back accuracy (which the IC has a tolerance of +/- 2 deg C) with my ATE. I'm still researching my solution and will (hopefully) post it here when I get one. Thanks all again.
Cristhian
(3,577 Views)
While the controller needs a thermocouple, you could use an RTD for the tester.
Lynn
(3,570 Views)