NI 9237 connection of external exitation

Solved!
Go to solution

Hi,

 

I am working with labview 8.6 and using a NI9237 mounted on a Chassis.

I am using 4 strain guauges with an exitation voltage of 10V for which i need to connect a external exitation source.

 

I am unable to figure out how to do this?There are 2 samll boxes with some connection terminals ,but I am not sure if i am doing it right.

Also, do i need to change my connections in anyway when I use an external exitation source?

 

 

0 Kudos
Message 1 of 6
(4,295 Views)

The NI 9237 Operating Instructions and Specifications document explains the connections in more detail.  Yes, the smaller connector on the bottom of the module is for external excitation (EX+ is on the top and EX- on the bottom).  When configuring your application, make sure to set the excitation source to be external and 10V.  If you wish to use 4 strain guages at 10V excitation, you are correct in using external excitation since the module can only provide 150mW (see page 15 of the previously linked document).

0 Kudos
Message 2 of 6
(4,294 Views)

Thanks.

I looked into the  document.

It does not seem to have any info on how to go about connecting those connectors.

I am having a problem with the hardware.

There is a black plug and some metal pieces which look like wire lugs,somewhat. 

I wanted to know how to push them into the slot and connect them.

0 Kudos
Message 3 of 6
(4,290 Views)

I think i have the external excitation set up now. Thanks for the info 🙂

 

I seem to have a problem still.

I am using the internal excitation now.

But irrespective of my excitation voltage the voltage  output seems to be constant.

 

If I measure the output from the load cells directly its varies with excitation voltage.

When i read the ouput via labview its independent.

 

Seems like black magic to me Smiley Surprised

 

Please help!!!

 

0 Kudos
Message 4 of 6
(4,258 Views)

Can you clarifywhat you mean by,

 

" If I measure the output from the load cells directly its varies with excitation voltage.

When i read the ouput via labview its independent."

 

What are your settings? What pins are you reading from? Can you provide a screenshot of your code?

 

Regards,

Glenn


Regards,
Glenn
0 Kudos
Message 5 of 6
(4,242 Views)
Solution
Accepted by karthikj85

Karthik85:

 

It is not black magic. The constant reading that you get is actually the mv/V value. That needs to be multiplied by the corresponding excitation voltage to get the mV reading. Read this article and hopefully it will clarify any black magic.

 http://digital.ni.com/public.nsf/allkb/21CA039D8F43C5188625729F005AC2F7

 

Best,

 

Santiago

Message 6 of 6