Multifunction DAQ
PT100 with NI6341 DAQ
I have a NI PXIe-1071 with a NI6341 multifunction DAQ.
I wante to use it with 4 PT100 temperature sensor.
Will it be possible (hardware/software  point of view) ? If not, does it exist a card able to get the acquisition with the PXIe 1071 ? (I saw the NI9216 but it is wotking only with CRIO)
Thank you for your advices.
(2,943 Views)
Accepted by topic author vincentlg
The 6341 is not a good choice for RTD temperature measurements.  As I recall (it's been a while), you need a board with a precision current source.   One such board is the PXIe-4357.  It's admittedly a pricey option for 4 sensors, so you may want to look into smaller dedicated RTD conditioners that attach via USB or LAN or something.
-Kevin P
(2,911 Views)
Just search for 'PT100 to 10V' converter,  they start at about 15$€ up to about 100, depending on the quality/housing/range...
Usually you use Pt100 to get accuracy...  so carefully look at the spec 🙂
Greetings from Germany
Henrik
LV since v3.1